This was indirectly tested in the
1994 ITE Book A #46:
The key says the answer is A. Bubbles cause overdamping, which doesn't affect MAP, so this is correct.
Underdamped systems underestimate (or accurately measure) diastolic pressure but provide accurate MAPs, so B & C are wrong.
Miller (6th ed p 1279-1280) recommends placing the transducer "5 cm below the sternal border in the fourth intercostal space" so E is wrong because zeroing at the sternal angle would underestimate the MAP.
If D is wrong (as the board claims), the implication is that long tubing either increases or doesn't change resonance in the system. Increased resonance = underdamping so what I'm taking away from this is that the board thinks (thought) that longer tubing causes underdamping. This is at odds to Hall and what intuitively and conceptually makes sense, but agrees with Big Blue.
I think that this old board question is wrong and may have been thrown out. Since much of Big Blue is based on old/remembered questions, I suspect this may be where Jensen got the bad info. I sent this off to Jensen using his errata form a few months ago but never got a response.
Idiopathic is right; long tubing contributes to dampening. Systolic pressure is underestimated; diastolic pressure is overestimated (or accurate); MAP is accurate.