Question regarding kVp and absorbed dose in radiography

This forum made possible through the generous support of SDN members, donors, and sponsors. Thank you.

radimaginggal

New Member
Joined
Nov 26, 2019
Messages
5
Reaction score
1
One resource says that increasing kVp increases patient dose as the result of more Compton scatter being produced.

Most other resources state that increasing kVp decreases patient dose because there is a more penetrating x-ray beam and less photoelectric effect.

What is your understanding of kVp and patient dose?

Members don't see this ad.
 
It only increases dose if you hold the mAs constant. If real life if you increase kVp you should decrease mAs which results in a lower dose. At least that’s my understanding of it.
 
Members don't see this ad :)
One resource says that increasing kVp increases patient dose as the result of more Compton scatter being produced.

Most other resources state that increasing kVp decreases patient dose because there is a more penetrating x-ray beam and less photoelectric effect.

To clarify the first statement, the amount of Compton scatter over the range of photon energies used in diagnostic radiology is essentially constant (independent of kVp). There is not more Compton scatter at higher kVp. It is because the probability of photoelectric effect interactions decreases with increasing energy that therefore Compton scatter accounts for a greater proportion of photon-matter interactions at higher kVp. Both Compton scatter and photoelectric effect contribute to patient dose. Because there is less photoelectric effect (in biologic tissues) at higher energies (at the higher end of the spectrum used in diagnostic radiology), the probability of energy absorption is decreased at higher kVp. Lower probability of absorption means lower patient dose.

However, the kVp not only affects the quality of photons but also quantity. Remember that kVp is a setting on your x-ray tube. It changes not only the energy of photons produced (quality; keV, which affects probability of energy absoprtion), but also the number of photons produced (quantity). More photons means more patient dose.

The balance of quality and quantity when you change kVp, and therefore the ultimate effect on patient dose, is explained in further detail in the Radiopaedia article above. The short answer is it's complicated in the real world because you will also adjust other parameters (mAs) in conjunction to achieve a certain image quality. As @SirGecko explained, if you increase only kVp and hold everything other parameter on the machine constant, dose increases.
 
Last edited:
"In general, lower kV settings will increase the dose to the patient compared to higher kV settings for the same imaging procedure and same body part, but the trade-off is that subject contrast is reduced with higher kV." pp 223 Bushberg Essential Physics of Medical Imaging
 
"In general, lower kV settings will increase the dose to the patient compared to higher kV settings for the same imaging procedure and same body part, but the trade-off is that subject contrast is reduced with higher kV." pp 223 Bushberg Essential Physics of Medical Imaging
I don’t understand what this quotation is referring to. This is in agreement with what @Cognovi wrote above.

The simple way I think about it is this:
Low kvp = mammo = more photoelectric = more soft tissue contrast
High kvp = chest = less photoelectric = less soft tissue contrast

Think about what a mammogram and the breasts on a chest X-ray look like.

And then remember that cxr is very low patient dose but mammo is relatively high.
 